I'm currently developing a Laboratory Information System which communicates with an Analyzer/Lab machine, so far I'm not having any problem with the communication from Analyzer to Computer (1-1) ratio. My problem is that I tried to connect multiple Analyzer into 1 Computer (connected to PC with USB-4 Ports RS232 Converter) and it messed up the results all results have been mixed up.
My question is how can I determined which Port sends a particular data so that I can separate and segregate it into each corresponding tables.
Here's my code in receiving:
Public Sub ReadData(ByVal sender As Object, ByVal COmPOrt As SerialPort)
COmPOrt = DirectCast(sender, SerialPort)
Dim received As String
RXString = COmPOrt.ReadExisting
received = Nothing
received = RXString
If (received = ack) Then
wh.Set()
ElseIf (received = enq) Then
Acknowledge(COmPOrt)
ElseIf (received = eot) Then
COmPOrt.DiscardInBuffer()
received = Nothing
Else
DataReceived = received(2).ToString()
With My.Settings
Select Case (DataReceived)
Case "H"
Acknowledge(COmPOrt)
Case "P"
PatientData1 = received.Split("|").GetValue(.PatientID).ToString().TrimEnd & vbCrLf
PatientData2 = received.Split("|").GetValue(.PatientName).ToString().TrimEnd & vbCrLf
PatientData2 = PatientData2.Replace("^", ", ").ToString().Trim
Acknowledge(COmPOrt)
Case "O"
OrderSampleID = received.Split("|").GetValue(.SampleID).ToString().TrimEnd & vbCrLf
OrderData1 = StrConv(OrderSampleID.Split("^").GetValue(0).ToString.Trim, VbStrConv.Uppercase)
FinalOrderID = OrderData1
Acknowledge(COmPOrt)
Me.Invoke(New MethodInvoker(AddressOf IdentifyOrder))
Case "R"
Data2 = Replace(received.Split("|").GetValue(.Parameter).ToString(), "^", " ").TrimStart() & vbCrLf
FinalData2 = Data2.Split(" "c)(0)
Data3 = received.Split("|").GetValue(.Measurement).ToString() & vbCrLf
Acknowledge(COmPOrt)
Me.Invoke(New MethodInvoker(AddressOf DisplayResult))
Case "Q"
Acknowledge(COmPOrt)
Case "C"
Acknowledge(COmPOrt)
Case "L"
Acknowledge(COmPOrt)
Me.Invoke(New MethodInvoker(AddressOf LoadResult))
End Select
End With
End If
End Sub
Public Sub OpenSP()
CloseSP()
For Each PortName As String In SerialPort.GetPortNames
Try
Dim NewSP As New SerialPort(PortName)
NewSP.BaudRate = 9600
NewSP.StopBits = 1
NewSP.DataBits = 8
NewSP.Handshake = Handshake.None
NewSP.Parity = Parity.None
NewSP.Open()
If NewSP.IsOpen Then
AddHandler NewSP.DataReceived, AddressOf frmSystemTray.COMPort_DataReceived
COMPort.Add(NewSP)
End If
Catch ex As Exception
MessageBox.Show(ex.Message, "Communication Error", MessageBoxButtons.OK, MessageBoxIcon.Exclamation)
Exit Sub
End Try
Next
End Sub
Below I provided the received data from Analyzer.
1H|\^&|||OPTI Medical OPTILION^2.00.0016^1601||||||Meas|P|2.2|2019011214426A
2P|1||ROSARIO|||||U17
3O|1||MEASUREMENT^1141||||||||||||Blood^Serum|||719815||||201901121442CA
4C|1|I|Na+ under 135 (Ref.Lim)^iCa under 1.15 (Ref.Lim)^|G11
5R|1|^^^Na^M^^|135|mmol/L|135 to 145|L||||||201803221320|OPTI Medical OPTILION^2.00.0016^1601F9
6R|2|^^^K^M^^|4.2|mmol/L|3.5 to 5.1|||||||201803221320|OPTI Medical OPTILION^2.00.0016^16013D
7R|3|^^^Cl^M^^|109|mmol/L|95 to 115|||||||201803221320|OPTI Medical OPTILION^2.00.0016^160184
0R|4|^^^iCa^M^^|1.13|mmol/L|1.15 to 1.33|L||||||201803221320|OPTI Medical OPTILION^2.00.0016^1601D6
1R|5|^^^pH^M^^|7.41||7.20 to 7.60|||||||201803221320|OPTI Medical OPTILION^2.00.0016^160116
2R|6|^^^nCa^C^^|1.13|mmol/L|||||||||OPTI Medical OPTI LION^2.00.0016^160184
3R|7|^^^AccNum^S^^||||||||||201803221320|OPTI Medical OPTI LION^2.00.0016^160120
4R|8|^^^Temp^S^^|37.0|?C||||||||201803221320|OPTI Medical OPTILION^2.00.0016^160184
5L|1|N08
And here's what I received when receiving from 2 analyzer simultaneously
1H|\^&|||OPTI Medical OPTI LION^2.00.0016^1601||||||Meas|P|2.2|20190112145069
2P|1||ROSARIO|||||U17
3O|1||MEASUREMENT^1141||||||||||||Blood^Serum|||719815||||201901121450C9
4C|1|I|Na+ under 135 (Ref.Lim)^iCa under 1.15 (Ref.Lim)^|G11
1H|\^&||||||||||P|E1394-97|2019011214512277
2P|1||212059||DALUSUNG, ALYSSA DAN||19960608|F32
5R|1|^^^Na^M^^|135|mmol/L|135 to 145|L||||||201803221320|OPTI Medical OPTI LION^2.00.0016^1601F9
3O|1|18106789130||||20180801011651|20180801001700||||||||1||||||||||F10
4R|1|^^^25^GluP|5.98|4||H||F|||20180801011719CF
5C|1|I|Flag^NORM_RANGEH|I63
6R|2|^^^K^M^^|4.2|mmol/L|3.5 to 5.1|||||||201803221320|OPTI Medical OPTI LION^2.00.0016^16013D
6R|2|^^^211^CREA3|72.35|6||||F|||20180801011655B9
7L|1|N0A
7R|3|^^^Cl^M^^|109|mmol/L|95 to 115|||||||201803221320|OPTI Medical OPTI LION^2.00.0016^160184
0R|4|^^^iCa^M^^|1.13|mmol/L|1.15 to 1.33|L||||||201803221320|OPTI Medical OPTI LION^2.00.0016^1601D6
1R|5|^^^pH^M^^|7.41||7.20 to 7.60|||||||201803221320|OPTI Medical OPTI LION^2.00.0016^160116
2R|6|^^^nCa^C^^|1.13|mmol/L|||||||||OPTI Medical OPTI LION^2.00.0016^160184
3R|7|^^^AccNum^S^^||||||||||201803221320|OPTI Medical OPTI LION^2.00.0016^160120
4R|8|^^^Temp^S^^|37.0|?C||||||||201803221320|OPTI Medical OPTI LION^2.00.0016^160184
5L|1|N08
Expected output should be:
1H|\^&|||OPTI Medical OPTI LION^2.00.0016^1601||||||Meas|P|2.2|2019011214526B
2P|1||ROSARIO|||||U17
3O|1||MEASUREMENT^1141||||||||||||Blood^Serum|||719815||||201901121452CB
4C|1|I|Na+ under 135 (Ref.Lim)^iCa under 1.15 (Ref.Lim)^|G11
5R|1|^^^Na^M^^|135|mmol/L|135 to 145|L||||||201803221320|OPTI Medical OPTI LION^2.00.0016^1601F9
6R|2|^^^K^M^^|4.2|mmol/L|3.5 to 5.1|||||||201803221320|OPTI Medical OPTI LION^2.00.0016^16013D
7R|3|^^^Cl^M^^|109|mmol/L|95 to 115|||||||201803221320|OPTI Medical OPTI LION^2.00.0016^160184
0R|4|^^^iCa^M^^|1.13|mmol/L|1.15 to 1.33|L||||||201803221320|OPTI Medical OPTI LION^2.00.0016^1601D6
1R|5|^^^pH^M^^|7.41||7.20 to 7.60|||||||201803221320|OPTI Medical OPTI LION^2.00.0016^160116
2R|6|^^^nCa^C^^|1.13|mmol/L|||||||||OPTI Medical OPTI LION^2.00.0016^160184
3R|7|^^^AccNum^S^^||||||||||201803221320|OPTI Medical OPTI LION^2.00.0016^160120
4R|8|^^^Temp^S^^|37.0|?C||||||||201803221320|OPTI Medical OPTI LION^2.00.0016^160184
5L|1|N08
1H|\^&||||||||||P|E1394-97|201901121453077C
2P|1||212059||DALUSUNG, ALYSSA DAN||19960608|F32
3O|1|18106789130||||20180801011651|20180801001700||||||||1||||||||||F10
4R|1|^^^25^GluP|5.98|4||H||F|||20180801011719CF
5C|1|I|Flag^NORM_RANGEH|I63
6R|2|^^^211^CREA3|72.35|6||||F|||20180801011655B9
7L|1|N0A
Finally, after many days of re-coding I got it to work as I wanted.
Here is my code on Multiple ComPort Receiving data from USB-4Ports RS232 Converter.
Dim Ports As New List(Of IO.Ports.SerialPort)
For x As Integer = 0 To My.Computer.Ports.SerialPortNames.Count - 1
Dim a As Integer = x
Dim PatientID(a) As String
Dim PatientName(a) As String
Dim OrderID(a) As String
Dim ResultName(a) As String
Dim ResultValue(a) As String
Ports.Add(New SerialPort)
Ports(a).PortName = My.Computer.Ports.SerialPortNames(a)
Ports(a).BaudRate = 9600
Ports(a).Open()
AddHandler Ports(a).DataReceived, Sub(obj As Object, evtArgs As SerialDataReceivedEventArgs)
Dim str(a) As String
str(a) = Ports(a).ReadExisting
If Not str(a) = "" Then
Dim received(a) As String
Dim DataReceived(a) As Char
Ports(a) = CType(obj, SerialPort)
Ports(a).RtsEnable = True
Ports(a).DtrEnable = True
Ports(a).ReadTimeout = 10000
received(a) = Nothing
received(a) = str(a)
If (received(a) = ack) Then
wh.Set()
ElseIf (received(a) = enq) Then
Acknowledge(Ports(a))
ElseIf (received(a) = eot) Then
Ports(a).DiscardInBuffer()
received(a) = Nothing
Else
DataReceived(a) = received(a)(2).ToString()
With My.Settings
Select Case (DataReceived(a))
Case "H"
Acknowledge(Ports(a))
Case "P"
PatientID(a) = received(a).Split("|").GetValue(.PatientID).ToString().Trim & vbCrLf
PatientName(a) = received(a).Split("|").GetValue(.PatientName).ToString().Trim & vbCrLf
PatientName(a) = PatientName(a).Replace("^", ", ").ToString().Trim
Acknowledge(Ports(a))
Case "O"
OrderID(a) = received(a).Split("|").GetValue(.SampleID).ToString().Trim & vbCrLf
OrderID(a) = StrConv(OrderID(a).Split("^").GetValue(0).ToString.Trim, VbStrConv.Uppercase)
Acknowledge(Ports(a))
Me.Invoke(Sub() IdentifyOrder(OrderID(a), PatientID(a), PatientName(a)))
Case "R"
ResultName(a) = Replace(received(a).Split("|").GetValue(.Parameter).ToString(), "^", " ").TrimStart() & vbCrLf
ResultName(a) = ResultName(a).Split(" "c)(0)
ResultValue(a) = received(a).Split("|").GetValue(.Measurement).ToString() & vbCrLf
Acknowledge(Ports(a))
Me.Invoke(Sub() DisplayResult(OrderID(a), ResultName(a), ResultValue(a)))
Case "Q"
Acknowledge(Ports(a))
Case "C"
Acknowledge(Ports(a))
Case "L"
Acknowledge(Ports(a))
Me.Invoke(New MethodInvoker(AddressOf LoadResult))
End Select
End With
End If
End If
End Sub
Next